Calibrating MPS Flow Sensor

Freitag, 11. November 2016, 22:43

I set up a little test loop and am trying to calibrate an MPS 400 Flow Sensor to a High Flow USB.

Looks like I may have been wasting my time the past hour or so.
I have aquasuite set to work in l/m, but that appears to be an invalid setting for calibrating an MPS Flow Sensor.
I can set the minimum to 1.5l/m, but the lowest the maximum field allows is 80l/m, when 7l/m is required.
Do i need to set aquasuite to l/h to calibrate the MPS Flow Sensor?

Mittwoch, 16. November 2016, 12:54

That menu will not work correctly when the unit is set to l/min. Please use l/h for the calibration. Might be fixed in a newer version (not the next one).

Mittwoch, 16. November 2016, 22:37

Thank you for update.